Calibration preview...
 
Share:
Notifications
Clear all

May 27 2026 APP 2.0.0-beta45 has been released !

Fully Multi-Threaded LNC, many improvements for the registration engine, platform upgrade, and further tuning of internal memory consumption and memory release back to OS.

Apr 14 2026: Google Pay, Apple Pay & WeChat Pay added as payment options

Update on the 2.0.0 release & the full manual

We are getting close to the 2.0.0 stable release and the full manual. The manual will soon become available on the website and also in PDF format. Both versions will be identical and once released, will start to follow the APP release cycle and thus will stay up-to-date to the latest APP version.

Once 2.0.0 is released, the price for APP will increase. Owner's license holders will not need to pay an upgrade fee to use 2.0.0, neither do Renter's license holders.

 

Calibration preview in 1.0.75 not working?

7 Posts
4 Users
0 Reactions
2,538 Views
(@cshine)
Molecular Cloud
Joined: 7 years ago
Posts: 3
Topic starter  

Hello, first time poster here.. is the calibration preview working properly in 1.0.75? I am probably doing something wrong as I don't see anyone else complaining but I can't get it to work.. I use it to verify that the flat frames are correcting properly before integration.

Thanks

1.0.75 Calibration
1.0.74 Calibration

 



   
ReplyQuote
(@astrogee)
Black Hole
Joined: 8 years ago
Posts: 234
 

I think the first image shows that your light frame did not have the calibration frames applied. The first column is missing MB, MD and MF that would indicate the bias, darks and flats calibration frames are applied. Why they weren't applied is another question.



   
ReplyQuote
(@cshine)
Molecular Cloud
Joined: 7 years ago
Posts: 3
Topic starter  

Yes indeed, I still haven't figured it out. It works perfectly in 1.074.1 with the same files. I don't even need to explicitly run the Calibration step, just load the files and view them, it's awesome. But 1.075 behaves very differently. I assume the rules for matching calibration to light frames are a little different now.

I re-ran the test in both versions with just 1 light frame, 1 flat and 1 bias. I've attached the console output when previewing the calibrated frame.

What's interesting is that when I explicitly run Calibration in 1.075 I see this:

16:13:59 - 2) CALIBRATE: assigning masters to light frames...
16:13:59 - 2) CALIBRATE: found master calibration frames for frame: 2019-10-21_21-52-36_-0.30_180.00s_200.00g_2x2bin_1.fits

 

But when I preview the light frame then I see this:

16:15:58 - IMAGE VIEWER: loading into Linear Calibrated Image Loader: D:\AstroProjects\19.10.18 M27 [180s 200 bin2]\Lights\Group1\19.10.22 M27\2019-10-21_21-52-36_-0.30_180.00s_200.00g_2x2bin_1.fits
16:15:58 - 2) CALIBRATE: VERIFY CALIBRATION MASTERS: all frames are verified
16:15:58 - FRAME DETAILS UPDATER: starting...
16:15:58 - FRAME DETAILS UPDATER: no new frames to add
16:15:58 - FRAME DETAILS UPDATER: rebuilding all frame details...
16:15:58 - FRAME DETAILS UPDATER: checking if frames were identified earlier...
16:15:58 - FRAME DETAILS UPDATER: adding frame marks...
16:15:58 - FRAME DETAILS UPDATER: updated succesfully
16:15:58 - 2) CALIBRATE: trying to find calibration details of light frame: D:\AstroProjects\19.10.18 M27 [180s 200 bin2]\Lights\Group1\19.10.22 M27\2019-10-21_21-52-36_-0.30_180.00s_200.00g_2x2bin_1.fits
16:15:58 - 2) CALIBRATE: DID NOT FIND calibration masters for light frame: D:\AstroProjects\19.10.18 M27 [180s 200 bin2]\Lights\Group1\19.10.22 M27\2019-10-21_21-52-36_-0.30_180.00s_200.00g_2x2bin_1.fits


This post was modified 7 years ago by Craig Shine

   
ReplyQuote
(@cshine)
Molecular Cloud
Joined: 7 years ago
Posts: 3

   
ReplyQuote
(@Anonymous 174)
Joined: 9 years ago
Posts: 5702
 

Interesting, I'll download them and have a look, thanks. Might it have to do with the light being binned x2? I'm not sure if the calibration files then match for the preview to work.



   
ReplyQuote
(@Anonymous 174)
Joined: 9 years ago
Posts: 5702
 

Only thing I see is that the calibration masters are 32-bit and your light is 16-bit. Tried that but also didn't work, so I'm now thinking a setting in the calibration data is wrong, I probably am overlooking something obvious so I'll add Mabula's eyes to this one. 😉



   
ReplyQuote
(@mabula-admin)
Universe Admin
Joined: 9 years ago
Posts: 5325
 

Hi @cshine, @astrogee, @vincent-mod,

The l-calibrated image viewer and calibration frames to light matching has not changed in 1.075 relative to 1.074.1.

Something else is causing this behaviour and it as the adjustment of APP's support for Bayer CFA sensors with uneven dimensions. Which is the case here.

What is happening is that for such a sensor, the masters created with APP 1.074.1 don't work in APP 1.075. Therefor the l-calibrated image viewer mode is not showing that the masters are applied to the light frames and so you will not see a calibrated image.

If you can create the masters again in 1.075 from the original calibration frames, please try that, then those masters should really work and you should see the l-calibrated preview 😉

If you no longer have the original calibration frames, please calibrate your frames in 1.074.1 and save them in 2) Calibrated. Then the calibrated frames are then fine for further processing in 1.075.

Apologies for this inconvenience, but as the changes are implemented now for better support for uneven Bayer CFA sensors, this incompatibility should no longer happen in future versions relative to APP 1.075.

Kind regards,

Mabula



   
ReplyQuote
Share: